T O P I C R E V I E W |
Shonky |
Posted - 09/21/2025 : 05:34:03 Hi I'm new here, great forum. I just bought my first geiger counter, a second hand GMC-300s from fleabay and there's something going on which doesn't make sense. I'm getting background readings mostly between 25 and 35 CPM but 0.00 on the microsieverts/h and the mR/h. Even with a handful of old thorium mantles, which got the thing clicking away quite merrily, there were only zeros from the dosage settings after an hour. I think it's a dud but just wanted to check whether I'm missing anything? Many thanks |
2 L A T E S T R E P L I E S (Newest First) |
3z_phoenix |
Posted - 09/22/2025 : 04:16:09 yes, i do RTFM :) pls. open main menue there u find "Init Setup" select this there u find "Factory Reset" pls. select this, it should load back all manufacturer defaults
check again with your radio-sources
if still stay the same problem over this "Factory Reset" u see "Calibration" for check the Values there not have to be Zero-Values...
in hardware description i found: High sensitivity tube M4011 installed (user can remove this tube install other tube if needed)
its the same Tube, like my GMC500+ use, i can give u my calibration values :) i only worry abit about "a User can remove this tube and install other" bcs. as u wrote your device had a pre-owner...
...hopefully we can get this device back into complete running...
|
3z_phoenix |
Posted - 09/22/2025 : 03:01:11 hi Shonky,
for me, this sounds like "lost" calibration-data, bcs. some math is used to convert the CPM-value into microSivert per hour.
please have a look into "initional settings", if you can find something like "Tube calibration" the value there must exist and no Zero-value.
sorry, i only have a GQ-GMC-500+ and i am not sure what Tube are used at your device... but sure our EmfDev will take-over so he read about your problem...
meanwhile i try to study the hardware details about your device...
kind regards, 3z_phoenix |
|
|