Hi I'm new here, great forum. I just bought my first geiger counter, a second hand GMC-300s from fleabay and there's something going on which doesn't make sense. I'm getting background readings mostly between 25 and 35 CPM but 0.00 on the microsieverts/h and the mR/h. Even with a handful of old thorium mantles, which got the thing clicking away quite merrily, there were only zeros from the dosage settings after an hour. I think it's a dud but just wanted to check whether I'm missing anything? Many thanks |
|
Reply #1
3z_phoenix

3 Posts |
Posted - 09/22/2025 : 03:01:11
|
hi Shonky,
for me, this sounds like "lost" calibration-data, bcs. some math is used to convert the CPM-value into microSivert per hour.
please have a look into "initional settings", if you can find something like "Tube calibration" the value there must exist and no Zero-value.
sorry, i only have a GQ-GMC-500+ and i am not sure what Tube are used at your device... but sure our EmfDev will take-over so he read about your problem...
meanwhile i try to study the hardware details about your device...