Posted - 06/22/2025 : 04:22:26
|
I have detector GQ GMC-500+, bought in August 2024. Few days ago, I got an issue with a device, that this started to show 0 CPM, 0.00 uSv/h, 0.000mR/h and null radiation with any settings - which actually cannot be so and there is clear, that something got wrong with the device. Battery is ok (we have also tested another similar battery) - in both cases the same result. We managed with device accurately. Visually this is like a new one. Also we have opened the backside and visually everything looks fine, all tubes are connected correctly, no any harm or damage. We have tried Turn off / turn on the device. Also tried to make Factory Reset. But these actions did not help. Tube1 Voltage displays 411 Volts (44%) Tube2 Voltage displays 41.33 %
What can be the problem? How can we diagnose - what is wrong? |
|
Reply #1
EmfDev
    
2349 Posts |
Posted - 06/24/2025 : 13:29:09
|
Open the device and try to tough the longer tube terminals with your finger and see if it can produce counts. |
 |
|
Reply #2
Raymoson

United Kingdom
1 Posts |
Posted - 07/27/2025 : 22:40:05
|
You should try resetting the device to default (if there is an option), then check if the firmware 1.12 is the latest. If the error persists, I think you should contact technical support, it may need hardware repair. In the meantime, just use the battery but pay attention to the battery life and backup data if necessary. |
 |
|
|