GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 GMC200 to 1-wire

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
rwcherry Posted - 12/02/2014 : 08:18:11
Has anyone interfaced the data output of the GMC200 to a 1-wire counter like the DS2423. It looks like the GMC200 only outputs 1.4V pulses and the DS2423 is looking for > 2.4V??

If anyone has done it, please share; otherwise I'll go ahead and do it.
1   L A T E S T    R E P L I E S    (Newest First)
ZLM Posted - 12/02/2014 : 11:07:23
The GMC-200 uses two diodes to clamp the audio output level to 1.4V. The original signal voltage is 9V.

You can remove two diodes and place a zener diode instead.

See:
http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=3407&SearchTerms=diodes