GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 uSv/h misure

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
ska Posted - 11/03/2012 : 07:20:39
Hi, I've a problem with my GMC-300 there is a singolarity in uSV/h misure at about 300CPM switch from 0.7uSv/h to 2.00 uSV/h, I reset to factory setting but the problem is still there.
Could I know how I can solve the situation?

Regards
3   L A T E S T    R E P L I E S    (Newest First)
ZLM Posted - 11/07/2012 : 01:18:53
The problem is the calibration data. In calibration menu, change your last uSv data to 3.75 or close value.
ska Posted - 11/05/2012 : 05:41:09
The problem is that at 240 CPM I have a 0.7 uSv/h as dose rate,
at 243 I have 1.91 uSv/h!

My SW version is 2.21

ZLM Posted - 11/04/2012 : 19:42:51
What is you exact problem? can you upload a picture?

What is your firmware version?