tomrdew |
Posted - 07/12/2015 : 06:35:23 running 320+V4 . works on data viewer ok. When try to use with data logger It does not work? running on win7 64bit. It will register but next time it loads shows unregistered. Logger Pro ver. 5.51. Is there a user guide for logger pro? Help... Thanks tom |
3 L A T E S T R E P L I E S (Newest First) |
ZLM |
Posted - 08/07/2015 : 08:28:00 That because Data Logger Pro is a generic software. Not only for GQ GMC320.
Use Data Viewer is free for all GQ GMC series Geiger Counter.
It does most of features as PRO. |
NaCl |
Posted - 08/07/2015 : 02:04:00 Data Logger Pro 12.95$
Awwwwwwww....
... So sad. |
ZLM |
Posted - 07/15/2015 : 13:13:03 The Data Logger PRO needs two settings for GMC-320+V4. 1. COM port. You can check the comport number from Data Viewer since it works. 2. Baud rate. The Baud rate must be 115200 in Data Logger PRO software.
Also, make sure your GMC-320+V4 set the baud rate to 115200 too. |