GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 GMC-600+ CPM should be divided by 2?

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
Sahim Posted - 08/17/2025 : 23:57:33
I have a GMC-600+ device.

1. Why divide and multiply by 2? For what purpose is this done?



2. Which is correct: the readings of the GMC-600+ device are incorrect but all other models are correct, or is it the other way around?



3. World map with real-time radiation https://www.gmcmap.com/ - should these indicators be divided by 2 or multiplied by 2?



4. The GMC-600+ device shows a level of 159 CPM in a room near ceramic tiles. What is the actual radiation level: 159 CPM or 80 CPM?



The device says: "159 CPM" -> "High". But according to the table 159/2=80 -> this is not "High" but "Average" level. So what is the level in reality, High or Average?


4   L A T E S T    R E P L I E S    (Newest First)
ullix Posted - 08/20/2025 : 00:37:21
quote:
The normal background in my garden in a private house outside the big cities is 33CPM = 0.0959uSv/h.


Perfectly ok; nothing unusual going on! (btw: nice, big garden :-))

Using a plastic sheet of 3.4mm thickness is what you would have done, had you read my "Potty Training"! It indicates that the radiation from the tiles is beta (though it does not proff it). Try double or triple the size and you will see further reduction of CPM.

Potassium is a natural compound radiating 90% and 10% gamma. Gamma has an energy of 1.46 MeV - among the highest of any natural compound - and is almost not absorbed by plastic up to several centimeter. What is stopped is beta.

So, I conclude that you do have beta radiation. Then it is INVALID to report dose as uSv/h, as the conversion of CPM to uSv/h is defined ONLY for Gamma, and, more precisely, only for Gamma originating from a Cs-137 source.

Overall, nothing to worry. Enjoy your sofa next to the tiles!
Sahim Posted - 08/19/2025 : 10:13:06
Thanks for the reply, ullix. My device has a Geiger tube LND 7317. I am not an expert in scientific papers. The normal background in my garden in a private house outside the big cities is 33CPM = 0.0959uSv/h.



This is 40 km from the city of Kyiv, Ukraine. The average value of the device close to the ceramic tile is 121CPM = 0.3422uSv/h. If you put a sheet of plastic 3.4 mm thick between the tile and the device, the readings will be: 56CPM = 0.1585uSv/h. I believe that the plastic delays beta radiation. I plan to buy a large sheet of plastic and install it between the sofa and the tile. I believe that the tile is radioactive. You write about potassium. What does this mean? The presence of potassium is not radiation? Maybe I don't need to install the plastic?
I still don't understand: the device reading "159CPM" - does it actually mean 80CPM or 320CPM? Should I multiply or divide the reading by 2?
ullix Posted - 08/19/2025 : 00:31:41
The GMC-600 counter has a Geiger tube with a sensitivity of 348 CPM/(uSv/h), while the other GMC counter use various Geiger tubes, all of them declared as having a sensitivity of 154 CPM/(uSv/h). That is roughly a 2x difference.

For more details see my article "GeigerLog-Calibration Guidance" here: https://sourceforge.net/projects/geigerlog/files/Articles/

I expect the normal background measured with a GMC-600 to be in the range CPM 30...50.

The higher background you measured at your tiles is probably due to the tiles containing Potassium; nothing unusual. And if so then you do see beta emission. To which the above sensitivities do NOT(!!!) apply!

For details see my article "GeigerLog-Potty Training for Your Geiger Counter" (available at the same location).

No Geiger counter can determine whether the tiles do have Potassium. For this you need a Gamma spectrometer; e.g. a RadiaCode device https://www.radiacode.com/.
Sahim Posted - 08/18/2025 : 00:27:13
For GMC-600+:



"159CPM" --> is 80CPM or 318CPM?


GQ Electronics Technical Support Forum © Copyright since 2011 Go To Top Of Page
Generated in 0.06 sec. Snitz's Forums 2000