and_kom

Ukraine
24 Posts

Posted - 10/27/2012 :  13:42:47  Show Profile  Reply with Quote
Is it possible to connect SI8-B Russian Pancake tube to GMC-300?

Image Insert:

140.22 KB

Young and radioactive

Edited by - and_kom on 12/28/2012 04:02:14
Reply #1

ZLM

1261 Posts

Posted - 10/31/2012 :  10:17:24  Show Profile  Reply with Quote
It should work. But you need to find a way to stick SI8B on back side of GMC-300.
Go to Top of Page
Reply #2

and_kom

Ukraine
24 Posts

Posted - 11/03/2012 :  22:44:05  Show Profile  Reply with Quote
I tried to connect but it showed very high reading of CPM which grew rapidly up to several tens of thousands...
Then I tried to connect it to another counter instead of SBM-20 and it seemed to work ok.
Seems GMC-300 circuit doesn't support such kind of device.

Young and radioactive
Go to Top of Page
Reply #3

ZLM

1261 Posts

Posted - 11/04/2012 :  19:34:45  Show Profile  Reply with Quote
You need to adjust the tube voltage.

Check your tube vltage.
Go to Top of Page
Reply #4

and_kom

Ukraine
24 Posts

Posted - 11/08/2012 :  10:53:22  Show Profile  Reply with Quote
quote:
Originally posted by ZLM

You need to adjust the tube voltage.

Check your tube vltage.



I've tried to adjust voltage but it didn't help. CPMs were growing up to 600-700 and more in each case and it was not expected behavior. Also I have noticed that speaker emitted more continuous clicks with less volume. Also I've tried to add 5,1 MOm anode resistor but got the same behavior.

Young and radioactive
Go to Top of Page
Reply #5

ZLM

1261 Posts

Posted - 11/15/2012 :  11:24:39  Show Profile  Reply with Quote
confrim your tube is good. Otherwise some other adjustmnent needed.
Go to Top of Page
Reply #6

and_kom

Ukraine
24 Posts

Posted - 12/28/2012 :  04:00:05  Show Profile  Reply with Quote
I confirm it works well with the DIY Geiger Kit (https://sites.google.com/site/diygeigercounter/home)

Young and radioactive

Edited by - and_kom on 12/28/2012 04:03:14
Go to Top of Page
Reply #7

hdripper

Switzerland
3 Posts

Posted - 03/26/2015 :  13:43:48  Show Profile  Reply with Quote
I'm sorry to be digging out this old thread, however I'm having exactly the same behavior with this tube, also in a GMC-300.

The background counts are way too high and easily exceed 600 CPM and often end up in the thousands. I am using an additional 22 MOhm resistor directly in front of the anode. Oddly, adjusting the tube voltage with R25 does not seem to have any significant effect on CPM.


EDIT: I have "manually" counted the actual CPM by recording the clicks through microphone and lookin at the waveform -> The clicks CPM from the tube is actually OK at roughly 150 CPM, as to be expected! So the problem would be related to how the signal is interpreted/counted by the mcu?

Is there anything I can do to the input circuitry to improve the impulse detection?

Edited by - hdripper on 03/26/2015 18:33:21
Go to Top of Page
Reply #8

Nudnik

40 Posts

Posted - 03/26/2015 :  23:48:42  Show Profile  Reply with Quote
Sounds like there is some noise on the wiring from the tube to the board. How long is the cable? Have you tried a shielded cable?
Maybe a shielded cable plus a lpf between cable and pcb would help.

Nudnik

under construction

Edited by - Nudnik on 03/26/2015 23:49:22
Go to Top of Page
Reply #9

hdripper

Switzerland
3 Posts

Posted - 03/27/2015 :  02:08:55  Show Profile  Reply with Quote
At the moment it is just a test setup with regular wires, albeit fairly short (10 cm). The proper shielded cable should arrive next week. Definitely something to try out!

On the other hand, since the original poster had exactly the same problem, which did not appear on the open source “DIYGeigerCounter”. So assuming that he was using a similar connection method on both counters, I’m not entirely sure if this will be solved by shielded cable, next week I should know more.

I still wonder why the audible clicks are correct but the display CPM reading is not... It seems like the clicks and CPM reading are each triggered seperately from the GM signal?

Also I am waiting for some 1 Gigaohm resistors to arrive so I can make proper measurements with the oscilloscope

EDIT: I searched some more through the forum and found this which looks very promising for now:

http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=3766&SearchTerms=cpm,increase

Edited by - hdripper on 03/27/2015 09:59:58
Go to Top of Page
Reply #10

ZLM

1261 Posts

Posted - 03/27/2015 :  15:22:10  Show Profile  Reply with Quote
I think there are some small noise signal along with the main pulses. The small noise can be picked up easily by the micro processor, but it is not easy by speaker and ear.
Go to Top of Page
Reply #11

Nudnik

40 Posts

Posted - 03/28/2015 :  04:30:11  Show Profile  Reply with Quote
Would be interesting to hook up a DSO to see whats going on.

under construction
Go to Top of Page